Hello, all. This isn't exactly a LabView issue (although I do use
LabView with my Agilent test eqipment). I have three Agilent test
instruments, an E5070B RF network analyzer, E4438C sig gen and an E4440
spectrum analyzer all connected on the same local area network (LAN).
When not requiring LabView, I connect to, and obtain data from, these
instruments using the Intuilink toolbars installed into MS Excel and MS
Word on a WinXP SP2 computer. Everything works fine as long as all 3
Agilent equipments are powered up. When I need only one instrument and
use the "connect to" toolbar item the "Searching busses" dialog box pops
up as usual but nothing else happens. After sniffing some IP packets on
the test instrument LAN I determined I had either an address resolution
protocol (ARP) or remote procedure call (RPC) problem in trying to
address the powered-off items. Putting static entries into the WinXP
ARP cache does not eliminate the RPC problem. Apparently the Intuilink
configuration cannot proceed (or perhaps the timeout is excessive) if
the configuration is trying to address some, but not all, powered-down
instruments. Does anyone know a work around other than having to power
up all three instruments? This is not a problem when using LabView.
Thanks for your time and ocmment. Sincerely,