09-04-2020 09:56 AM
Hi. I tried to change the currnet 1.0A to 1.5A not a 1.0 -> 0 -> 1.5A
I think it works normally when voltage does not exceed 6V, however if the voltage over 6volts then, It doesn't work.
1. Initialize -> false (when the reset process added, then the current go to zero ampere) -> if the voltage doesn't exceed 6V it works. (1A to 1.5A)
2. Initialize -> True (when the reset process added, then the current go to zero ampere so I can't use this option) -> if the voltage exceed 6V it works.
3. Initialize -> False (the reset process was do not work, then the current move from 1A to 1.5A) -> the voltage exceed 6V. This shown error message.
I need 3... Can anywon tell me how?
Solved! Go to Solution.
09-04-2020 09:59 AM
the error message
09-08-2020 02:59 PM
If you are doing this in LabVIEW post your code. Most likely you are setting the current range to 1.0A in code, perhaps without knowingly doing it by leaving out a terminal connection. Then, when you try to set the output to 1.5A, but the max range is 1.0A you will get an error. Initializing or Reseting to default values might get you the 3A range, which would allow 1.0A or 1.5A to be set just fine. Later operations might change things but we need to peek at the code to be sure.
Craig
09-08-2020 09:29 PM
Hi, First of all thank your kind answer!
Yeah I think so that can be the solution to my question.
actually, I tried to change the default current compliance(Probably now is 1 A) 1A to 1.5A
but, I don't know how to change it.
I tried to change source range 0 to 3 (from labview) 1 is small 2 is 0.5A 3 is 1 A but there is anyother options to choose 1.5A or 3A (b2901A max current is 3A)
09-10-2020 02:53 PM
Try something like this...
This should allow you to see if the Current Range or Voltage Compliance is your issue.
09-12-2020 03:31 AM
Thanks for kind answer.
1. possible
It works, only that needed voltage didn't exceed 6V
2. impossible
I need 1.5 ~ 1.8 A but, the needed voltage exceed 6V so it doesn't work.
3. Copliance voltage to 8V
It shows the error message.
(I know that B2901A limit votlage and current were 200V, 3 A.)
Where am I programming wrong?
09-14-2020 02:22 PM
That meter has a voltage/current sourcing limit. In your "possible.png" your measuring 5.99V while sourcing 1.5A. A small increase in current will increase the voltage, but since the voltage is out of range the meter will throw an error. See the manual diagram below. Best you can do is 3A/<6V or 1.5A/<21V. You can't go 1.8A at >6V. You will need a different SMU for that.
09-14-2020 11:12 PM
Thank you so much.
1.8A was approved without going through the lab view... It seems that it was outputting that it was not approved.