I have a PXI 4065 under LV 8.5.1.
When a make a current measurement without the circuit under test the DMM measures 1mA. i don't know how this is possible or if this is normal.
This affects me beacause i have a circuit under test where the spected measurement is around 1mA so i don't know if the DMM acquired the correct measurement.
Any ideas? Regards
11-04-2008 07:25 AM
Hi ethan85,
Thanks for the forum post and I hope your well today.
I noticed your post has been left unanswered for a few days now.
For example, if we left the input to a DAQ device and did a voltage reading, the value will float - and sometimes this can be noise or sometimes it can rail up/down to the range specified for the task.
What is the range for your task set too?
I guess the best way to see if this value is ok, is to test your device using a high current, and see if it still returns to 1mA.
Also with DAQ devices they make current measurements by measuring the voltage across a shunt resistor. This could well be the cause of the 1mA.
My last thought, is its usually not advisable to leave an input floating - so maybe some forum of pull down resistor could be used when the device DUT isn't connected.
Please let me know what you think,
Kind Regards,
11-04-2008 09:12 AM
Hi Hillman
Thanks for your response.
I corrected this problem, I modified the range from 0.5A to 0.1A and the measurement on the floating point was of 0.17mA so this works for me.
I'll work in a application for dynamically changing the range of the DMM.
Regards