The gain adjust value is the end result of the shunt calibration and is used to scale strain measurements on this channel. Normally, this value is close to 1, which implies that the measured value and simulated value are equal.
You can find out this value by referencing the column labeled "Gain Adj Val" in the Calibration window. You can also retrieve this value programmatically in LabVIEW by placing a DAQmx channel property node and selecting Analog Input>>General Properties>>Signal Conditioning>>Bridge>>Shunt Cal>>Gain Adjustment.
Also, please refer to the following knowledge base information on how voltage readings on this device can be interpreted.
http://digital.ni.com/public.nsf/websearch/21CA039D8F43C5188625729F005AC2F7?OpenDocument
Let me know if this clears everything up!