DAQ Assitant: Setting "signal input range" outside calibration range
I have a question concerning the calibration and measurement of thermocouples using the DAQ assistant.
I want to measure temperatures at cryogenic temperatures (liquid nitrogen, ~-196°C).
To calibrate the thermocouples, I used a two point calibration (ice water, 0°C and liquid nitrogen, -196°C).
Now I cannot set my "signal input" minimum below -196°C, this causes an issue since now I cannot measure below -196°C. Especially I cannot see the "negative" (below -196°C) noise of the signal when in liquid nitrogen.
I can set the minimum signal input to a lower temperature if I disable the calibration, however than the measurement is highly inaccurate.
The error i receive is Error -200077 (which is explained here https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000000P7gySAC).
Is there an option to let LabVIEW extrapolate the form the calibration range?
I couldn't find anything online so I appreciate any help. Thank you in advance for your answers.
(3,263 Views)
Hi Max,
where (and how) do you apply "calibration" data in your DAQAssistent?
I wouldn't apply scaling to a thermocouple reading - atleast not within DAQmx.
If needed I would apply the scaling on the temperature reading I get from DAQmx in a 2nd step in the VI:
The constant in the snippet is suitable for a perfect sensor without measurement errors, but you can apply any (and how many) coefficients you like…
(3,224 Views)
Hi Gerd,
Thank you for your reply.
To answer your question.
Where : I apply my callibration data/ calibrate directly in the DAQ assistant. The same approach just in NI MAX is described here https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z000000P80ZSAS .
How : I place my sensors in icewater and liquidnitrogen of which I know the temperatures (0 and -196°C)
So I do no generate a DAQmx code from the DAQ assistant and alter it.
I would have assumed, that Labview has full dataset (e.g. https://www.thermocoupleinfo.com/type-k-thermocouple.htm ) of each thermocouple type (in my case K type) and uses the two calibration point to fit the characteristic thermocouple type curve to each specific sensor. The mV/°C-slope of the thermocouple types is known, there should be only small differences/offsets due to the junction right?
I understand your suggestions as follows: Use the internal calibration/ NI factory setting for the thermocouple type and read out the temperature. In a second step apply a scalling fucntion to calculate the actual temperature. Is that correct?
(3,211 Views)
Hi Max,
So your assumption is maybe wrong? 😄
When you apply a new calibration you are replacing the known conversion curve - that's my assumption…
Yes. Either do as I have depicted above - or follow the last sentence in the KB entry you linked to:
This type of calibration can be done programmatically in LabVIEW by taking measurements at a known reference level and creating a custom scale. For more information on custom scales, please see the Related Links section below.
(3,207 Views)