DaqMX timing with DI and DO tasks

TimingAndStartTasks.png

I have a question about applying sample clocks to multiple tasks, as I keep getting errors. I know this is a fundamental issue that I am misunderstanding, so I appreciate everyone's help and suggestions.

 

I am using multiple CDaq chasis with multiple AI, Thermocouple and DIO modules.

I have created 4 tasks:

 

1 for all AI

1 for all TC

1 for all DI

1 for all DO

 

When I run the code pictured above, it works. When I try to apply timing to the DI and DO modules, though, I get an error that the hardware conflicts. I applied continuous samples with the same rate that is wired into the AI and TC tasks.

 

I've researched into the issue and tried some of the examples, but I am still a bit baffled.

 

I want to avoid having an uncontrolled data rate from my DI and DO modules which could overwhelm the system.

 

How can I assign timing to those tasks?

Also, should I fundamentally change the framework I am using for the tasks?

 

Thanks for all the help.

 

 

 

0 Kudos
Message 1 of 18
(737 Views)

What cDAQ chassis do you use?
https://www.ni.com/en/support/documentation/supplemental/18/number-of-concurrent-tasks-on-a-compactd...

 

Have you tried a very basic example that generate/acquire hardware timed task per type (AI, AO, DI, DO) for all types?

Santhosh
Soliton Technologies

New to the forum? Please read community guidelines and how to ask smart questions

Only two ways to appreciate someone who spent their free time to reply/answer your question - give them Kudos or mark their reply as the answer/solution.

Finding it hard to source NI hardware? Try NI Trading Post
0 Kudos
Message 2 of 18
(713 Views)

Hi -

I appreciate your help!

 

It is a cDaq 9189.

Hardware Timed Tasks

# of Tasks per Chassis

# of Tasks per Module

Analog Input

3*

1

Analog Output

1

1

Counter Input/Output

4, 6

4, 1

Digital Input

1

1

Digital Output

1

1

 

I am using 2 chasis's.

I am trying to use 2 DIO modules (both are 9403).

Both of these modules are on the same chasis.

 

Should I change the timing source on the sample time settings?

I tried that a bit but still got the error - but maybe I missed something.

 

GOB34_0-1724899437397.png

 

0 Kudos
Message 3 of 18
(706 Views)

Hi GOB,

 


@GOB34 wrote:

I am trying to use 2 DIO modules (both are 9403).

Both of these modules are on the same chasis.


The table shows "1 DI task per chassis" and "1 DO task per chassis".

And it says "only one DI/DO task per module".

Those are the rules.

 

Unfortunately all the details of your implementation is hidden in a subVI - and you don't provide information on that subVI…

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 4 of 18
(671 Views)

Thanks for the reply @GerdW.

 

 I understand the limitation noted in the table, the thing is it works when I don’t include the timing for the DI and DO tasks - so i don’t see how it is an issue with number of tasks but more I am messing up some timing setting.

 

If I don’t specify the timing for a DI or DO module, and I am using a producer/consumer QMH structure like the “continuous measurement and logging(daqmx)” template, then what rate is it running at?

 

Regarding any sub VI - there is nothing else going on besides the tasks get initiated (with false and then the task references brought in here.

 

Appreciate the help!

 

0 Kudos
Message 5 of 18
(664 Views)

Hi GOB,

 


@GOB34 wrote:

I understand the limitation noted in the table, the thing is it works when I don’t include the timing for the DI and DO tasks - so i don’t see how it is an issue with number of tasks but more I am messing up some timing setting.


The timing engine is the limiting part in here: this engine only supports a limited number of tasks…

 


@GOB34 wrote:

If I don’t specify the timing for a DI or DO module, and I am using a producer/consumer QMH structure like the “continuous measurement and logging(daqmx)” template, then what rate is it running at?


It will run at the rate of the measurement/producer loop: when you don't set a timing in the DAQmx task then you should set a timing in your producer loop…

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
Message 6 of 18
(659 Views)

Could I make the DI and DO one task?

0 Kudos
Message 7 of 18
(642 Views)

Hi GOB,

 


@GOB34 wrote:

Could I make the DI and DO one task?


How would you distinguish between inputs and outputs when using DAQmxRead/DAQmxWrite on your task?

No, you cannot mix inputs and outputs within one task…

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 8 of 18
(640 Views)

OK, so in the future with this hardware, I will need to separate DI and DO onto different modules in order to get hardware timing. I can't do that for my current project.

 

That is somewhat limiting and frustrating but I understand now!

 

Also annoying because it effects the rest of the system timing wise -

 

i.e. I am acquiring from all tasks in the same loop - all the other input modules have timing set via hardware.

 

Now I have this "rogue" other DI module that does not have hardware timing possible. So I have to introduce a wait because of that module.

0 Kudos
Message 9 of 18
(618 Views)

Hi GOB,

 


@GOB34 wrote:

it effects the rest of the system timing wise - i.e. I am acquiring from all tasks in the same loop


NOBODY enforces you to put all DAQmx code into just one loop!

I usually use one loop per task…

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
Message 10 of 18