Differential Mode in DAQ M series board
Hello all!
I have a doubt concerning the Differential Mode in DAQ M series board. In the board's manual is saiing that the voltage limit is -10 to 10V, but somebody knows if this limit is always referenced to ground, or in differential mode is comparing with the +8 channel (ex. for channel 0 - voltage diference between ch0 and ch8).
Thanks in advance, best regards.
Paulo Carmo
(3,886 Views)
In attach you can find the schematic.
(3,885 Views)
The voltage limit is always referenced to ground, so the voltage limit is between + and GND and - and GND. If you want wire more voltage, you need to use high-pass filterto reduce the common mode.
(3,862 Views)
Paulo,
differential mode means that you measure the difference between two input channels. The channels used for this on E- and M-series boards is n and n+8. The difference between the channels might go up to +- 10V.  So there is no ground reference in this type of measurement
However, the device has a limited capacity against overvoltage. So if the offset of either of the channels succees a certain value in regard to the ground of the measurement system (the gnd of your PC), you could possibly damage your hardware.
Therefore it is recommend to use "BIAS-resistors" in such cases (or if you simply don't know the offset).
You can find a very good image on possible ) recommended connections in your DAQ-help. Open the DAQmx Help and navigate to DAQmx Help >> Measurement Fundamentals >> Signals >> Analog >> Connecting Analog Input Signals >> Measurement System Types. Just activate that entry (should have a "book" icon) and in the right side, you should see a something like a reference poster. I suggest you to print it out....
hope this helps,
Norbert
[EDIT]: corrected a typo
Message Edited by Norbert B on 02-26-2009 08:03 AM
Norbert
Expert: Geometry
(3,856 Views)
Paulo,
that kind of connection is possible, if you can guarantee that the offset of your lines is not exceeding a critical value. The critical value is defined by the power on the lines. So it might be possible that the connection in the schematic already could damage the M-series device. Therefore, consider using BIAS or isolation (please note that there are isolated AI M-series devices available). Make sure reading the specs of the device that your device is capable of managing that input.
hope this helps,
Norbert
Norbert
Expert: Geometry
(3,844 Views)
Hi Norbert!
Do you know what is that limit? The over-voltage protection? My voltage will not excedd 30Vdc.
Regards and thanks for your fast reply.
Paulo Carmo
(3,840 Views)
Paulo,
no, i don't know what device you are using therefore i don't know the specs.
To be on the save side, just use BIAS, the smaller the value, the greater the measurement error intrudced by them but the better you save the M-series device......but you'd possibly can calculate that error.
hope this helps,
Norbert
Norbert
Expert: Geometry
(3,834 Views)