Hello NI Forum Members,
I have been trying to communicate with the MKS946, first with NI Max. However, I always get a Timeout Error when I query its serial number '@253AD?;FF' or any command. Can anybody see what is wrong with the command?
Regards,
RL
Solved! Go to Solution.
01-22-2024 08:28 PM
To figure out the problem, first, we need the manual of this device with the communication well documented, which you still need to provide. The manufacturer requires registration to access the manual.
https://www.mks.com/f/946-vacuum-system-controller
01-23-2024 09:04 AM
Thanks for the reply @Santhosh. The manual is freely available online, here for example:
MKS 946 SERIES OPERATION AND MAINTENANCE MANUAL Pdf Download | ManualsLib
The instructions for remote control of the manual start on page 94.
01-27-2024 05:59 PM
I have been able to communicate with the MKS946. My mistake was that I was using a null modem cable instead of a straight one.
Regards,
Ricardo