Thank you to Ben & David for your inputs, most helpful.
Natalie
dhlabview@yahoo.com (David Huff) wrote in message news:...
> natwilkie@hotmail.com (Natalie) wrote in message news:...
> > Hi all,
> >
> > I have an SCXI-1303 terminal block plugged in to an SCXI-1102 module
> > that I'm wiring up for just straight voltage inputs, I'm using
> > shielded pair wiring for each channel, my question is do I need to
> > connect the shield to the Earth Ground Solder Lug from each of the
> > cables?
> >
> > If this is the case do I also connect the other end of the shield to
> > the instrument I am measuring, or just one or the other?
> >
> > Thanks for your input
>
>
> > Natalie
>
> You should connect the shield to the ground in the terminal block, but
> not to the ground on the instrument, because connecting both grounds
> to the shield might cause a ground loop.
> Check out the following weblink at NI, it has a great introduction to
> proper and improper wiring techniques.
> <>
> note that the entire link is enclosed in brackets<>
>