Hello,
Thank you very much for using our products. I have checked with the documents and found that the SIT 3.0 or 3.0.1 do not support cRIO-9201. This is the reason for not being able to use the calibration VIs for cRIO-9201. NI certainly would like to add the support for cRIO-9201 with SIT, but at this moment I am not sure when that will be.
By using the property nodes of FPGA , you can acquire the LSB weight and offset information for each channel in cRIO-9201. You can pass these informations to a cluster, and use these calibration data to the Real Time VI to acquire correct readings from your channel. Please see the attached Word file , where the VIs for FPGA and RT are showed. Ignore the parts in the VIs for analog output.
Thank you.
Tanvir
National Instruments, Japan
10-19-2006 10:07 PM
Hello Tanvir,
Thanks for your valuable response, I got LSB weight, offset and other calibration data from actual cRIO-9201 as you mentioned, and changed calibration VIs. At result, cRIO-9201 still acquired incomplete data.
In first, the data in which cRIO-9201 acquired using cRIO-9215 calibration VIs are nonlinear for actual voltage values. By contrast, using LSB weight and offset from cRIO-9201, acquired data by cRIO-9201 are linear for actual voltage value but not equal to actual voltage values.
Perhaps Using LSB weight and offset for cRIO-9201 cannot calibrate A-D converted data completely. Anyway, I can apply these data because acquired data with cRIO-9201 are linear for actual data values. (The linear data is easy to fit actual data)