01-04-2007 03:31 AM
More info.
If I go into Measurement & Automation Explorer my devices are all there.
If I do a Self-Test I get the 'Success' box: 'The device has passed the self test'.
If I do a Reset Device I also get the 'Success' box: 'The device has been reset successfully'.
But then if I go into a Test Panel in the Finite and Continuous mode nothing happens al all, no signals appear. In the On Demand mode I get the following error:
Error -200474 occurred at Test Panel
Possible Reason(s):
Specified operation did not complete, because the specified timeout expired.
Device: PXI1Slot3
Any more suggestions?
Cheers,
Sarah
01-04-2007 06:35 AM
01-04-2007 08:55 AM
Yay, it's working now! 🙂
Un-installed DAQmx 8.3 then reinstalled it again. Also it looks like my first PXI-6133 card is faulty, now I have replaced that too all is well.
Thanks for all your help! Much appreciated.
Sarah