Hi,
I am getting an error no. 201398 while calibrating a strain gage. Could anyone help with this?
Thanks
Piyush
08-10-2011 11:10 PM
Can you provide some more details? How are you calibrating this gage? Is it in MAX or in a LabVIEW program? Was there any text provided with this error code?
08-11-2011 03:08 PM
I am using LabVIEW program. I have 10 strain gages, and 9 out of 10 could be calibrated. But, one showed the error 201398. Attached file shows the error description.
Thanks
Piyush
08-11-2011 04:16 PM
Something must be wrong with that gage or its connections.
It is basically telling you that the correction it is trying to make is too big for the daq device to handle. It looks like it might be the null offset value, but it is unclear from your screenshot.
What kind of gages are you using? Are they set up as 1/4 bridge, 1/2 bridge, or full bridge? It is reading a value that is so far out from where it should be that it can't set the property within the range of the DAQ device.
Try measuring millivolts across the bridge with a multimeter.