Im using labview 2011 to measure voltage readings from a 34970A instrument. Im able to measure the voltage, write it to a file and plot waveforms.
Im also able to let the user change the resolution. When i try to control the NPLC(Number of Power Line Cycles), the VI runs prefectly, but im getting errors in the instrument. The error codes are 521 & 103. 521 is a communication error : Input Buffer Overflow and 103 is "Invalid separator".
I have attached the sub-VI that configures the voltage and gets the values from the instrument. I noticed that, if i run my Initialize.vi in normal mode, it shows error, but if i highlight the executio, it runs normally. Is it something to do with the timeout values? The other sub-VIs works perfectly.
Thank you,
Eureka
(7,083 Views)
Accepted by topic author Eureka
Since you are resetting the instrument, try giving some delay after the initialize function.
(7,082 Views)
Dennis,
It works!!! you're amazing!!! I added a time delay right after the initialize.vi as you advised.
Thank you very much Dennis,
Thank you,
Eureka
(7,076 Views)