The max in min values include the custom scale, so calculate what the new_min would be using the custom scale and the raw_min and set that to be the minimum of the range. Then do the same for the max. Your range should be good to go then!
Kristen H.
08-19-2009 04:16 PM
08-19-2009 04:23 PM
Squinn21:
Regardless of scaling errors, I believe the 6024 can only read voltage inputs, not current. You would need to place a 250 ohm precision resistor across the DAQ input terminals to convert the 4-20mA into a 1-5V siganl the DAQ can read.