Where I can find some examples on how to use NI-9775 with a cDAQ chassis (not cRIO).
I can't find them in Example Finder; it can be that I must search them using a different key..
04-17-2020 07:43 AM
Looks like this module is only for c RIO.
The NI 9775, a 4-channel digitizer, can measure transient phenomenon like faults in electricaltransmission lines from lightning strikes or structural failure events at 20 MS/s/ch. Themodule's store and forward architecture allows up to 128 Mbits of measurement data to be sentback to the controller and analyzed. The module has a built-in analog reference trigger, or youcan use CompactRIO and LabVIEW FPGA to develop an advanced trigger based on low-speed streaming data for added flexibility.
04-17-2020
09:27 AM
- last edited on
04-04-2025
01:39 PM
by
Content Cleaner
I'm not sure about it, for a couple of reasons:
Where did you find that info?
Thanks
12-01-2021 01:39 PM
Did you ever get an answer to either of your questions (either the original or the follow-up)? We are looking to use the NI-9775 to repeatedly capture high speed data and save that data to a file for post-processing. I'm hoping there is an example of someone doing this out there. Did you ever get your project started?
12-02-2021 01:46 AM
Hi @EricWilliamsEPT,
I've never received any answer
09-23-2022 05:17 PM - edited 09-23-2022 05:21 PM
There are examples installed with the CompactRIO and drivers. Does not install in the 64-bit version of LabVIEW. But if you open the 32-bit version and look for 9775 you should find:
The path is Toolkits and Modules->FPGA->CompactRio->Analog Input.
Just registered that you have a cDAQ chassis. Let me check if I can find anything.
09-23-2022 07:03 PM
Same idea as the cRIO, but in the DAQmx section of the help: Browse Task.
Hardware Input and output -> DAQmx -> Analog Input....
Analog Input - Filtering.vi calls out NI 9775 support.
I did not have a module but created a simulated cDAQ-9185 added the NI 9775 and ran it in continuous. Seems to match the documentation on streaming 4MS/s max rate in continuous mode.
In finite mode all channels can be captured at 20MS/s/ch but the buffer is shared so you are limited to a total of 8.38861MS/s for 1 ch and 8.38861 / 4 MS/s .
Attached my modifications to the standard example.