Fluctuation in torque measurement using NI 9237
Sorry to ask a question that is hopefully pointless, but how did you calculate the weight -> torque values? Are you sure you used the appropriate arm length when applying the weights during calibration, and carried out the correct conversions between e.g. kg mass and Nm torque?
GCentral
Message 11 of 17
(1,072 Views)
Hi Patel,
@patelbrijesh535 wrote:
I am sorry about manufacturer manual, this is very old torque meter and we don't have manual available for specific model.
I am using 10V internal excitation from NI 9237 module.
Did you connect the NI9237 and that sensor correctly?
Did you verify the 10V excitation voltage?
Can you cross-check with a "known working" (and probably not so old) sensor to verify the correct cabling and behaviour of the NI9237?
What else did you do to verify your measurement setup?
Can you even supply the complete sensor type/name?
Message 12 of 17
(1,072 Views)
I have identified the wires as per its old connection. when it was installed with old data acquisition system. my model is TP - 5KMCB from KYOWA, I contacted them and they have provided attached manual, which doesn't cover my model.
Message 13 of 17
(1,068 Views)
I was mind full of weight. I did use arm length 0.5m and applied weights from 0.5 kg to 9 kg.
Message 14 of 17
(1,065 Views)
Hi patel,
then why don't you follow that manual?
It recommends excitation voltage in the range of 1…4V!
Message 15 of 17
(1,063 Views)
This manual doesn't have model which I am using, so I decided to go as per pervious connection where 10V was used.
Message 16 of 17
(1,062 Views)
When you did the calibration, what output did you get with 0.5 kg and what at 9 kg?
You said this is an old torquemeter.  It is based on slip rings.  Are you sure it isn't damaged and getting the electrical signals across the slip rings?
Have you tried using an ohmmeter across the the various bridge leads to see what resistances it is giving you?
Message 17 of 17
(1,050 Views)