Hi,
I used a Fluke45 to read a AC voltage. The program works well in an old computer however, it shows an error when I use a new PC to control it. The error is attached. I doubt if it is the bus interface problem since when I opened the program, it also shows there are warnings. I attached both warnings and program also. Thanks.
05-16-2012 05:09 PM
Hi hxw,
The error you are seeing is a timeout error:
“Error -1073807339 occurred at an unidentified location
Possible reason(s):
VISA: (Hex 0xBFFF0015) Timeout expired before operation completed.”
There are many reasons why you could be seeing this error including a bad connection to your device, but you may want to start here:
05-16-2012 05:19 PM
Not sure why you posted those warnings. Don't you understand what they mean?
In any case, with a GPIB connection, check the cable because it could be a failure in the handshaking. You might also want to change the GPIB bus timing in MAX.