09-05-2006 11:23 AM
Hi,
I am generating an executable file for a client using a developer computer (laptop). However, the executable file on the client’s laptop is running fine except one channel is not reading the same. The input is .216 vol but reading .170.
I am measuring voltage and temperature (8 input and 2 temperature). I am using 6062E card. I did a self-calibration in MAX but still reading wrong.
I would appreciate your help.
09-05-2006 04:54 PM
09-06-2006 10:48 AM
Thank you for your reply Rob,
Shorting the channels did not improve the reading. It seems all the channels are not reading correctly not one as i stated before. However, discconect the user laptop from the power outlet (120 v) and running it on battery I get the correct reading. I connect all the 8 AIGND together I improved the reading %70.
I am measuring input voltage as NRS from a0 to a3 and a8 to a11. a5 and a6 as thermocouple k type. generated 5 and 0 V from DAC0OUT and DAQ1OUT.
I welcome Any more suggestions.
09-06-2006 11:21 AM