Getting only half wave during data acqusition
Hello all,
              When I acquire data using cDAQ NI 9215 and chassis 9171, my data is getting tripped/hatched by the lower side ( I have attached a file for your reference). And this happens with some channels at different times while other channels give good output. Kindly, help with the problem with this. What can be done to rectify this problem?
Figure 1: tripped/hatched signal
Figure 2: tripped/hatched signal
Figure 3: Estimated result.
Note: I also checked with NI MAX Test Panel, I am getting similar results.
Message 1 of 10
(3,421 Views)
Are these signals floating?  Or the grounds are not referenced to each other?  That is what this looks like.
You might want to have a look at this article: Field Wiring and Noise Considerations for Analog Signals
Message 2 of 10
(3,403 Views)
Most likely (IMHO) a missing Input Bias current path.   Solutions in the link crossrulz provided...   (one or two 1Meg resistors from the anlag - input (or both)  to AGND usually solve that.
Greetings from Germany
Henrik
LV since v3.1
Message 3 of 10
(3,385 Views)
I have noticed similar results when the vertical range is not correct.
=== Engineer Ambiguously ===
Message 4 of 10
(3,360 Views)
I am using NI 9215 and Current Transformer sensor connected vai BNC cable.
Message 5 of 10
(3,341 Views)
Can you tell me how to correct this?
Message 6 of 10
(3,338 Views)
Your sensor is likely floating.  This means it needs something to pull it towards ground.  Added 1MOhm resistors on each side of your transformer to AIGND should do it.
Message 7 of 10
(3,311 Views)
Thankyou for your answer. Can you guide me how to ground this using resistor. I am using current transformer and a BNC cable for this.
Message 8 of 10
(3,288 Views)
Which type of 9215 do you have? The BNC input version should have a 100kOhm Resistor to AGND/COM already ...
Which version are you using?
For the screw or spring version try a 1M Ohm resistor between AI- and COM terminal
If you use the 9215 BNC version and you run into that problem, please provide  a schematic of all your cabling. Do you have a handheld multimeter  (DMM)  around? can you check the voltages at the BNC shield .. between the shields and to PE/Ground ?
I don't have a 9215 (BNC) around, but I assume you should measue a resistance of 200kOhm with the DMM from one BNC connector shield to the next channel BNC connector shield. (Just in case one 100k is blown in history)
Henrik
LV since v3.1
Message 9 of 10
(3,278 Views)
I am using 9215 (BNC).
Message 10 of 10
(3,201 Views)