HP 83652A connection problem

Hi,

         I'm using LabVIEW 2009 and have installed NI device drivers(NI-488.2).I've connected HP 83752A synthesized sweeper to the PC through GPIB cable.The GPIB address of this instrument is automatically showing up in the VISA open function.The problem is in the Agilent I/O connection expert,under Instrument I/O on this PC this instrument address is showing up with a Failed symbolbefore it saying there was  a communication failure.Even If I send *IDN? string I'm getting "There was a communication failurae as response".can any one help me out 

Regards
RobinHood
0 Kudos
Message 1 of 4
(2,920 Views)
Why are you using the Agilent utility? Since you have an NI controller, you should use MAX.
0 Kudos
Message 2 of 4
(2,916 Views)

thank you fo your reply.can you please tell me what to do in Max

 

Regards
RobinHood
0 Kudos
Message 3 of 4
(2,913 Views)

To do what? Just communicate to the instrument? If so, right click on the instrument listed there and select 'Communicate with Instrument'? You can also select Tools>NI-VISA>VISA Interactive Control.

 

p.s. Going to the Help menu is also a good idea.

 

Edit - you must also install NI-VISA.

Message Edited by Dennis Knutson on 03-09-2010 06:13 AM