Help Using Labview to control Kikusui power supply with a PIA4850 controller

HI I need some help using Labview to control 4 Kikusui power supplies. We have a PIA4850 Kikusui power supply controller. I can use the visa test panel and see the PIA4850 but can see any of the power supplies. I have very limited use of Labview and only wrote a couple programs using GPIB. The PIA4850 is a USB controller and can control up to 32 PS. I can get it to work with a Excel demo software that came with the controllers so I know that it is working. Any help will be appreciated. I am using LabView 8.2 and have updated to the Ni-DAQ 9.0. I have a program that I need to incorporate the PIA4850 into once I figure out how to operate it. Thanks in advance.

0 Kudos
Message 1 of 5
(5,165 Views)

hi antique,

  You can download labview drivers for your controller from

http://www.kikusui.co.jp/en/download/en/archive1.html#PIA4800series

 

 

 

Thanks and regards,

srikrishnaNF

 

Regards,
Srikrishna


0 Kudos
Message 2 of 5
(5,161 Views)
There are also the Kikusui ki4800 Power Supply instrument drivers. This collection of VIs should allow you to control your Power Supply.
- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
Ben Sisney
FlexRIO V&V Engineer
National Instruments
- - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - - -
0 Kudos
Message 3 of 5
(5,139 Views)

HI and thanks for the help. I did get one power supply to work but have another problem. I need to control 4 power supplies at the same time and can only operate the one that is addressed N5!C1. If I set the other to N6!C1 and so on I get an error stating

 “Error -1074135008 occurred at ki4800_2 Configure Voltage Level.vi Possible reason(s):IVI Error in the method "get_Item" in the component "Kikusui4800".

E_IVI_UNKNOWN_CHANNEL_NAME: Unknown channel name (N6!C1)”

 

I am assuming that I should be able to change the node number and that will change the supply that I need to use. Everything works great if I use N5!C1 I can change any of the supplies to address 5 and can control it with no problem. I can even have multiple supplies with the same address N5!C1 (don’t think I should do that) and change them all at once. I need independent control of each supply.

 

I am in a pinch here and need to figure it out but could use some help. Any ideas will be greatly appreciated.

 

Thanks

GBlair
Download All
0 Kudos
Message 4 of 5
(5,117 Views)

OK I figured it out. I just overlooked a part in the Kikusui IVI-COM instrument driver programming guide. After installation there is a scan utility that needed to be ran when all of the power supplies are connected? I ran it when I had the first unit connected but had not ran it with all of then connected. Once that was done they were all recognized and worked great. Thanks for the help.