There are many reasons why you can have a noisy signal. Signal conditioning may or may not be the answer for you, but it is the first place to look. Typically, temperature measurements should not be that noisy.
What is your voltage range? Is your DAQ setup for that range (if input level is programmable)? What temperature (voltage) swings are you looking at?
There is a lot of useful information on techniques for minimizing noise in the Field Wiring guide.
Signal conditioning offers a few different techniques for improving signal quality including amplification and filtering. There are many resources available to learn about using signal conditioning for different applications including the following: Improved Signal Quality via Conditioning and Increase Your Measurement Accuracy by 10X with Signal Conditioning.
Are you using the PT 100 probe from Ohmega? If so, this is what they have to say:
"Wiring for the PR-11-2-100-1/16-6-E (3-wire 385 Pt 100 ohm sensor element) Probe:
One RED & two BLACK wires; the RTD sensor element is across the RED wire & either of the BLACK leads; the two black wires are connected to the same point on the RTD sensor and used for lead wire resistance compensation. At room temp the RTD will measure approx 108 ohms across the RED & BLACK leads, and measure a short across the two black leads. Beyond this, I cannot help you with the connections to the National Instruments. However, an RTD is a resistive element (resistance is directly proportional to temp increase) and does not require COLD JUNCTION COMPENSATION like a thermocouple."