A 4 wire measurement would be best, the meter will maintain constant
voltage/current through the component and help isolate the other
components from the measurement.
When using autorange, many meters will return a transitional value; check that it's not an autorange issue by measuring twice and looking at the second value returned. I generally configure
any meter for the nominal value before measuring. Some meters will take a long time to autorange and you may get a VISA timeout waiting for the instrument to return a large value indicating an open.
The choices are
Config meter for the expected value (manual ranging)
Trigger, wait and measure with autoranging
Autorange and measuring twice (discarding the first value)