09-02-2008 08:42 AM
I fully agree with Dennis on this one. This is a hardware issue. You do want to configure the DMMs so one triggers the other. However, this will produce a delay from one to the other, but it should be highly repeatable. This allows you do take some pretrigger points on the slave DMM to compensate. This does not, however, ensure that the data points are collected at the same time (the clock phases on the DMMs may or may not be synchronized). If you can share the clocks, in some fashion, between the DMMs, you should do this, as well. If you can't, you should be able to determine the offset and do a resampling to get things back where they belong (this is not a really desired method of doing things, as resampling can miss things like sudden data spikes).
Note that we could give more help if we knew the brand and model of DMMs you are trying to synchronize.
Good Luck! Let us know if you need more help.
09-02-2008 09:04 PM
Hi Gray
Thanks for your comments. Actually I use two Agilent DMMs (3458A). Ya it is a harware problem. When we use external timing signal to sample triggring and Trig Event, we cannot find fix relation of Trig Event between two DMMs. Master slave combination can find fixed relationship with repeatable data, but sampling rate is controlled by programme. Alreardy I did the synchronization with fixed time delay using second option. But still I have prolem. Some time programme works, but some time the 2nd DMM not trigger. I attached progrrame code here. Can you give some comments to fix this problem problem.
Wije
09-03-2008 07:28 AM
09-03-2008 08:51 AM