Initial decay increasing with sampling rate

Solved!
Go to solution

Hello everyone,

 

I'm experiencing an initial decay in my signal that increases with increasing sampling rates. I've included my acquisition code and examples of acquiring the signal with 1 kHZ and 10 kHz sampling rates.

 

I'm acquiring three physical channels using a USB-6003 DAQ board. By the time I run my code, the amplifier is already working and sending data to another computer through a RS232 cable.

 

I'd love to know what could be causing this, since the code is very simple. Could it be a hardware issue?

 

Thanks in advance,

 

Lucas

Download All
0 Kudos
Message 1 of 10
(3,835 Views)

There can be several causes, some explained here: http://digital.ni.com/public.nsf/allkb/72FEC45B7EAA8E2586257E26006E6E21

0 Kudos
Message 2 of 10
(3,798 Views)

Hello lkmaciel,

Is the issue occur when you acquiring one physical channel?

Message 3 of 10
(3,780 Views)

I tried that now, and it doesn't occur. Only if I acquired a different channel before.

 

For example, acquiring 3 channels simultaneously causes the initial decay in all three.

If I then acquire only channel 1, I have a decay on the first run but not in any subsequent run.

If I try to acquire channel 2 then, again it will show decay only on the first run.

 

So this is only happening when I run my VI for the first time with the channel.

0 Kudos
Message 4 of 10
(3,777 Views)

What amplifier are you using?  How did you wire up the signals?  Is the signal ground connected to AI_GND?


GCentral
There are only two ways to tell somebody thanks: Kudos and Marked Solutions
Unofficial Forum Rules and Guidelines
"Not that we are sufficient in ourselves to claim anything as coming from us, but our sufficiency is from God" - 2 Corinthians 3:5
0 Kudos
Message 5 of 10
(3,747 Views)

I'm using a Kistler 5073 charge amplifier (https://www.kistler.com/?type=669&fid=61429&model=document).

 

The wires go straight from the amplifier to the DAQ board (ground and channels 0, 1 and 2).

0 Kudos
Message 6 of 10
(3,741 Views)
Solution
Accepted by topic author lkmaciel

In your code example you leave "input terminal configuration" terminal unwired. For USB-6003 default value for channels 0-4 is "differential". So if you connect wires as RSE this configuration can lead some issues in your measurements

Message 7 of 10
(3,718 Views)

@M_Stas wrote:

In your code example you leave "input terminal configuration" terminal unwired. For USB-6003 default value for channels 0-4 is "differential". So if you connect wires as RSE this configuration can lead some issues in your measurements


Hmm, interesting point. But actually then how the OP can even see signals measured, if the terminal is set as Differential (as default if this is correct)? The pinout is totally different....

0 Kudos
Message 8 of 10
(3,716 Views)

Thanks, M_Stas! Changing it to RSE solved the problem. I just need to get my head around understanding it now.

0 Kudos
Message 9 of 10
(3,710 Views)

@Blokk wrote:

@M_Stas wrote:

In your code example you leave "input terminal configuration" terminal unwired. For USB-6003 default value for channels 0-4 is "differential". So if you connect wires as RSE this configuration can lead some issues in your measurements


Hmm, interesting point. But actually then how the OP can even see signals measured, if the terminal is set as Differential (as default if this is correct)? The pinout is totally different....


When you have a floating input, who knows what is left on that channel.  It most likely would trail the positive input or even stay at or near ground.