Hi,
I am trying to read data from a node with my DAQ usb-6251. I've also connected the node to an oscilloscope. Whenever DAQ samples data from the node, the oscilloscope shows that the voltage shows peaking. I am wondering why this occurs because it seems to corrupt the correct voltage on the node.
07-12-2010 11:27 PM
Input impedance!
Your scope most likely presents a load of 1-10MOhm (or 50 Ohms depending on your setting but I think we can assume you've got the 1-10Mohm option selected ) The USB DAQ device is a "Student Aid" quality and has an input impedance of @ 140K Ohms. When both devices are connected together you WILL see a change on the O'scope when the input is gated to the DAC inside the DAQ device. Basically, you just re-discovered Hiesenberg's "Uncertainty Principle" and found out for yourself that the ACT of observation MUST change the reACTion you wish to observe.