Issues with communicating with two channels of a Keithley 2220-30-1 Dual Channel Power Supply

We use a VI program to communicate with a Keithley 2220-30-1 Dual Channel DC Power Supply. When we try to communicate with both channels simultaniously, an error message randomly appears:

 

Error -1073807360 occured at an unidentified location

Possible reason(s)

VISA: (HEX 0xBFFF0000) Unknown system error (miscellaneous error)

 

This originates from the Keithley VISA driver. When I use another VI program to communicate with only 1 channel on the power supply, the program runs perfectly, using the exact same VISA drivers in the block diagram. I have tried many things to solve this issue, with no success. I have attached the VI code on this post to give clarification to my problem. Where would be the best place to start to fix this?

 

Thanks, dglock18 

0 Kudos
Message 1 of 7
(3,520 Views)

change the VISA reads and writes to "Synchronous"  Right-Click>>Synchronous I/O Mode>> Synchronous.  the watchband glyphs are a dead givaway that "Asynchronous" is selected.

 

Keithley may eventually learn to properly buffer commands.


"Should be" isn't "Is" -Jay
0 Kudos
Message 2 of 7
(3,514 Views)

Niether the instrument or the GPIB standard supports simultaneous communication. GPIB is and will always be a sequential communication bus. Jeff''s recomendation forces that on your program but it should be rewritten to reflect reality.

0 Kudos
Message 3 of 7
(3,511 Views)

@Dennis_Knutson wrote:

Niether the instrument or the GPIB standard supports simultaneous communication. GPIB is and will always be a sequential communication bus. Jeff''s recomendation forces that on your program but it should be rewritten to reflect reality.


Keithley's are notorious for having commands with different priorities.   Getting two nearly simultaneous "Immediate" commands is something their FW designers could have handled "more elegantly."


"Should be" isn't "Is" -Jay
0 Kudos
Message 4 of 7
(3,503 Views)

@Dennis_Knutson - I understand, and thank you. I figured that was the case, but are you saying I should rethink my logic entirely for communicating with the Keithley? How would I be able to communicate with different two channels another way? I am fairly new to the Labview community... I'm sure this is a relatively easy fix

0 Kudos
Message 5 of 7
(3,495 Views)

Thanks Jeff! I'll try it and report back to see if that resolves the issue entirely. 

0 Kudos
Message 6 of 7
(3,494 Views)
You could use the error in/error out wires to enforce dataflow. Don't have communication structures that execute in parallel. Use a single Action Engine for all communication. Use VISA Locks. All different possibilities.
0 Kudos
Message 7 of 7