05-29-2025 07:24 AM
I'm trying to use E4980 LCR meter, after using many others LCR meters, for example more advanced E4990.
But E4980 stuck my VISA libraries, when i try to read.
Sometimes I get unexplicable errors like -1073807360 & -1073807240.
But most of the times the read instructions take me to a stucked situation that i solve only turning off the LCR.
Very strange that MAX test panel seems to work with no problems, until I try to open a VISA session with Labview.
The instrument by itself has seen by the pc control panel as an IVI instrument, is this correct?
05-29-2025 10:17 AM
I found that the problem is inside the init routine.
I replaced the init routine included in the driver of E4980 with the one of the E4280, now it's working.
I tried to exclude CLEAR and option of VISA Open, but I don't solved the problem.