LabView Thermocouple Temperature Readings
I'm using SCXI-1001 with SCXI 1102 and SCXI 1303 (LabView 8.2) to record temperatures from k-type thermocouples.  My problem is that the temperatures I'm seeing on LabView are off by approximately 5-10 degC AND they drift +/-10 degC throughout the day.  From what I've read, I'm thinking it might be related to CJC.  Right now I just have CJC Source set to "Constant" with a value of "25".  I can make adjustments to this to help with the offset, but the temperatures just drift throughout the day and the readings become incorrect again.  The room temperature does change quite a bit in the area where I have this setup, and the temperature drift seems to be related to that.  Any help or suggestions would be much appreciated. Thank You.
(5,732 Views)
Yes, it is related to using a constant cold junction temperature value and the actual temperature of the junctions changing throughout the day.
Set your CJC to use the onboard thermistor rather than constant within your task.
(5,724 Views)
I have tried changing the CJC source from "Constant" to "Built-In" and the following error message comes up:
"Measurements: CJC Source cannot be set to Built-In for the
specified thermocouple channel.  The physical channel does
not support built-in CJC temperature sensor.
If your hardware contains a CJC temperature sensor on the
physical channel corresponding to the built-in CJC source,
make sure that the hardware configuration (including and accessories
and/or terminal blocks) is correct.  Alternately, specify a different
CJC Source, or use hardware with a built in CJC temperature
sensor."
(5,692 Views)
Accepted by topic author Zawer
Hi Zawer,
Your SCXI-1303 has a high-accuracy thermistor cold-junction temperature sensor, so the "Built-In" setting should be supported.  Try resetting the 1102 in Measurement and Automation Explorer and then restarting LabVIEW.  Is this error still being thrown?
Sara Lewandroski
Applications Engineer | National Instruments
(5,650 Views)
It works now.  The problem was that the 1102 was not properly configured in Measurement and Automation Explorer.  Thanks for the help.
(5,626 Views)
I reset my 1102 pragmatically every time my applications start. I have had a few issues where the temperature will be "stuck" at a strange value a little above ambient upon startup, but since I started resetting all the time - no problems. I think it also helps with initialization, especially when people hit abort rather than doing a proper shutdown the last time the program was used - leaving the 1102 in an unknown state.
The reset certainly doesn't hurt anything, and is a piece of cake with the DAQmx Reset Device vi.
edit: Smiley Very Happy Check out the funny spell check change.
Message Edited by Broken Arrow on 04-16-2010 03:28 PM
Richard
(5,607 Views)