02-24-2011 08:24 AM
02-24-2011 03:49 PM - edited 02-24-2011 03:51 PM
A lot of the speed is going to be driven by the resolution of the frequency measurement. For instance: if you have a 10Khz sample and you set the resolution to 0.1 hz the measurement must take at least 10 seconds to complete (1/0.1) It looks like, from your example, that you are using the autorange function and the graph is centrered @30KHz so, 100,000 hz range at 5.5 digits would be 5Hz resolution and take 1/2 sec. The same measurement a 4.5 digis would complete in 50mSec.
02-25-2011 12:37 AM
Hello there
Thanks alot for reply.
Sorry I did not understand your reply fully. Where can I reduce frequency resolution. As I am not expert in Labview and I just modified the old Labview program. So where can I set resolution to like 1Hz so that I can take more than 2 measurements in a second.
Thanks again
03-07-2011 07:39 AM
Hi faheemkhan,
Is still an issue for you?
Because this device is from a 3rd party, you must contact the producer of the device with this type of question.
03-14-2011 08:22 AM
Hi Kristoffer
I managed to solve the issue. Actualyl in the default program, which I donwloaded from NI website, there was some unnecessary commands which were making whole devices slowe. So now I just use Fetch Measurement.vi and put the device on Auto Arming and this makes the device at maximum speed.
Thanks
Faheem
Denmark