Lock-In-Amplifier SR830
Hi, I have an SR830 lock-in amp but instead of the standard GPIB or RS232, I would like to get the signal via BNC cable through an NI 9234 box. I was able to get some numbers with the DAQ Assistant, but I don't think it is getting the same data as the front panel has shown. Are there some settings that I am missing? Thank you in advance for helping!
(3,550 Views)
I suggest you look at the manual, it is well documented; what I am saying here is only what  I recall and my not be entirely correct.
I assume you are using the outputs from the channels to digitize. These outputs are ±10V and were meant to drive a strip chart recorder. 10 V represents full scale for your measurement, that is, if the the sensitivity is set to 1V and you get 10V out you have a 1V signal. So in order to get the same value as the display you need to keep track of the sensitivity scale and scale your values.
Lastly, I am not sure of the update rate for this output, I think is kind of slow, ie, no more than 512Sa/s, it may depend on the time constant . So what I am trying to say, adjust your digitizing rate accordingly.
mcduff
(3,523 Views)
Thank you!!! It is more complicated than we thought originally. Will go back to GPIB or RS232 on the instrument.
(3,503 Views)