Monochromator interfacing and programming
I am trying to interface a monochromator and write a program. The monochromator is from a local company. They provided initialization parameters and code to communicate. The communication was done through RS232 via a USB converter. I find the interfacing was successful (see a.png). I tried to write a program based on the code provided.
Code:
Initialization Parameters for Serial Communication
Baud Rate: 19200 bps
Byte Size: 8
Parity: None
Stop Bits: 1
Selecting a wavelength
  1. Send 5 to controller and wait for the acknowledgement 5.
  2. Send 10; ack 10.
  3. Send WNByte1; ack WNByte1
  4. Send WNByte0; ack WNByte0.
  5. Wait for 20.  The monochromator will send this code after the specified wave number is set.  If you want to stop the process before receiving this code, send 100.  Also, if you are receiving any other code here instead of 20, it means an error condition.
WNByte1 = Integer value of (Wave Number / 256).
WNByte0 = Remainder of (Wave Number / 256).
However, I did not receive an acknowledgment from the instrument while using LabVIEW program (see b.png). Even in VISA test panel, I did not receive any acknowledgment. I am using LabVIEW 2010 on a Windows 10 run laptop. Could anyone please help me find the issue?
Thank you.
Pradeep.
(3,759 Views)
Hi pclab,
don't use BytesAtPort, it's wrong 99.9% of all time…
(3,729 Views)
@GerdW wrote:
Hi pclab,
And in the off chance that the instrument does not have a termination character you should put some wait between your write and your bytes at port. Your bytes at port is likely getting 0, so you're reading 0 bytes.
Another thing that you need to verify is whether your protocol requires you to send a termination character. Since you said that it doesn't work in the test panel I suspect that the instrument may be sitting there waiting for a termination character.
(3,712 Views)
@johntrich1971 wrote:
@GerdW wrote:
Hi pclab,
Also in this case get rid of BytesAtPort and simply wire a "1" to have VISARead wait for 1 byte. No need for another wait function…
Mind to attach that code? Most often things like termchars are recognizable…
(3,709 Views)
@GerdW wrote:
Hi pclab,
As usual you are correct GerdW (at least for the cases that we're given - if there are other cases where the ACK might be different then this would not work). It's been a while since I've worked without a termination character - and when I did I often had responses of varying length (though I could and would usually calculate the expected response).
Also, if they provided code to communicate why are you trying to reinvent the wheel? Why not use the provided code?
(3,690 Views)