Multiplicity Error, too many data values of ...
i believe the error is because of the limit line i drawn according to the values in the msg , but the whole program runs well, except the error message. And i dont know how to clear this error.
equipment is Agilent E4440A Spectrum Analyzer
(1,906 Views)
settled
(1,900 Views)
Hi noob,
it really would help to provide a (real) solution instead of marking the word "settled" as solution…
What have you done to solve that Agilent-internal error?
(1,883 Views)
Accepted by topic author Noob_intern
Sorry about that,
So for this error, means there is too many duplicate data values at the same "position".
As for my program, there is a part where it will draw a limit line using 13 points.
So When i test run the program the first time, the limit lines will be created once already. the limit lines data will be saved in the instrument.
and when i test run the program after modification the 2nd time, 3rd time, 4th time.. more and more same limits line are created.
Thus causing this error~
The solution: find commands from manual, to delete all limits lines at the start of the program. So everytime there will only be 1 set of limit lines in the instrument
instrument: Agilent E4440A Spectrum Analyzer connect via GPIB
BR,
An intern w 1 mth labview experience so far
(1,865 Views)