My serial communication commands dont work in MAX
I am trying to read and writ a command to a temperature oven. The commands offered by NI-MAX don't work.
I attached a picture with a sample command that the manual gave. This device uses GPIB.
Message 1 of 16
(4,176 Views)
Does it need a termination character?  (probably)  Did you send a termination character?  (I don't see one in your command.)
Hit enter at the end of version.  Or type /n or /r.
Message 2 of 16
(4,167 Views)
Thanks for the response. Yes I tried \n and \r.
Message 3 of 16
(4,143 Views)
I also am not sure which GPIB to choose.
I tried GPIB0 and GPIB4 and GPIB5 then "Scan for Instrument".
Not sure.
Message 4 of 16
(4,141 Views)
Leave it as GPIB0.  You should only need to change that if you have multiple GPIB controllers on your system.
Message 5 of 16
(4,135 Views)
What model of oven and/or controller are you trying to talk to?  There are plenty of other settings to worry about that might show up in the manual.
Message 6 of 16
(4,133 Views)
That looks like it's supposed to be a query.  Did you click the "Query" button?  (Alternatively, did you click "Write" followed by "Read"?)
Bill
CLD
(Mid-Level minion.)
Message 7 of 16
(4,123 Views)
Thank you for the response on GPIB0.
Here is the manual.
Message 8 of 16
(4,076 Views)
Thank you for the response.
Yes I tried both of these options. Write and read also query
Message 9 of 16
(4,073 Views)
How many ovens do you have connected?  I see 10 instruments listed in your NI-MAX list.  Are they all powered on?  How long are the cables between them?  Have you specifically set the address on each oven?  What is the address you are attempting to contact?  (In your first post its GPIB0::4::INSTR, but its not shown in the list in the latest image.)
GPIB has some restrictions on cable length (2m between instruments/repeaters) and a total of 20m of cable length.  Also I think its 2/3 of the instruments have to powered on at any time.  If you have a huge long run on cabling, and most of the instruments are off your feeble GPIB querry just gets lost and eventually you get that timeout error.
I would try to send either "Status?\r\n" or "Temp?\r\n".
Craig
Message 10 of 16
(4,053 Views)