NI 9214 Temperature Calibratoin
I have a NI 9214 DAQ that I'd like to perform a temperature calibration on.  I noticed that if my temperature range is from 0 to 100C, the SW requires me to test both bounds of the temperature range.  However, my temperature standards are 0 and 95C.  Is there a way that I can get the software to accept that my 2 point calibration is not at the bounds of my measurement temperature range and to just extrapolate out the 2 point calibration curve?  I'm using the example code for continuous thermocouple DAQ measurements.
How do I configure a 2 point calibration from 0-95C for a range of 0-100C?  If I do this in DAQmX, will this still work in my VI (which doesn't use express VIs)?  Or do I have to programmatically apply the 2 point calibration to the output data?
(2,754 Views)
Hi sonun,
I wanted to ask you some clarifying questions.
1. Are you trying to do an external calibration of you device?
2. Are you trying to create a table of prescale/postscale values?
3. What type of thermocouple are you using?
Also, any screenshots of your code are helpful.
Taylor
(2,700 Views)
Hi
1. Are you trying to do an external calibration of you device? Yes, I am placing the thermocouples in controlled temperature settings (water bath + ice) and measuring the difference.
2. Are you trying to create a table of prescale/postscale values?  Yes. but I would like to do this for a full load of 16 thermocouples
3. What type of thermocouple are you using?  Type T
(2,680 Views)
Hey sonun,
It seems as though you're wanting to create a custom scale for your application.  I'm including a snippet of a possible solution that you could implement in your code.  I'm a little confused though since the DAQmx VIs that you're using should account for the T type thermocouple that you're using.  Take a look at the image and try creating a custom scale that will represent your 0 to 95 degree C range.
Taylor
(2,652 Views)