01-29-2014 03:13 AM
Dear Sir
I am using NI PCI 6281 which has analog input range of -10 v to 10 v. My signal has 10.2 V AC, can I connect this with DAQ card safely or it will damage the port?
Waiting for prompt reply.
Kind Regards
Muhammad Irfan
01-29-2014
04:39 AM
- last edited on
08-15-2024
03:59 PM
by
Content Cleaner
Hi Muhammad,
read the specifications: it allows at max ±11V against AIGND in working operation. The inputs are protected with ±15V/±25V depending on device off/on state!
01-29-2014 04:50 AM
Dear Sir
The signal is in differential mode (i.e two wires one + and other -). No GND being used in the connection scheme. So should I go for connections or not?
Kind Regards
Muhammad Irfan
01-29-2014
11:00 AM
- last edited on
08-15-2024
04:00 PM
by
Content Cleaner
tronoh wrote:
The signal is in differential mode (i.e two wires one + and other -). No GND being used in the connection scheme. So should I go for connections or not?
Are those signals referenced to ground somehow? Even in differential mode, each input should not go beyond the +-11V to ground.
You might want to give this white paper a read:Field Wiring and Noise Considerations for Analog Signals
01-29-2014
11:44 AM
- last edited on
08-15-2024
04:00 PM
by
Content Cleaner
@crossrulz wrote:
tronoh wrote:
The signal is in differential mode (i.e two wires one + and other -). No GND being used in the connection scheme. So should I go for connections or not?
Are those signals referenced to ground somehow? Even in differential mode, each input should not go beyond the +-11V to ground.
You might want to give this white paper a read:Field Wiring and Noise Considerations for Analog Signals
At this point we have no way to judge if the signal you are attempting to connect will damage the system.
Let's use some examples and discuss the real use of that third or fourth wire on your primary power cord.
Assumtions: Something in your system is tied to a publically available power grid ( or, You may be on this planet "Earth", interplanataray travellers are refered to their respective species representatives for advisement)
there are exactly 2 options:
Attaching the two systems together will provide a common point of reference between the two of no greater than 20Volts before damage will occur to the DAQ device and potentially the computer housing it.