NI USB-6009 Problem: 1.4V measured at open input with DAQ Assistant
I have just chekced it:
I connected potential deviders and what I measure is a mess:
Potential devider with two resistors between 5V and Ground. R1 = 100k always (at the 5V side), R2 variable
R2 = 1M    ==> Voltage should be 4.54 V, I've measured 3.4 V
R2 = 100k ==> Voltage should be 2.5V, measured 2.3 V
R2 = 10k  ==> Voltage should be 0.45 V, measured 0.50 V
So what "This behavior is normal and does not affect the measurement when a signal is connected"???
I'm really upset!
Message Edited by johanneshoer on 06-05-2009 08:58 AM
Greetings Johannes
Using LabVIEW 7.1 and 2009 recently
Message 11 of 13
(992 Views)
The analog input impedance is specified as 144 kΩ.  What you measured is consistent with that.  The USB-6008 is not suitable for measuring signals with high source impedance.
I agree that this is rather low input impedance for a voltage measuring instrument, but at least NI does specify it.
If you need to measure a voltage with a high source impedance, you either need to buffer the signal with a low output impedance amplifier or select a different measuring instrument.
Did you read the specs before you bought the USB-6008?
Lynn
Message 12 of 13
(981 Views)
Of course I did not! Who reads specs? 🙂
Yes I did, but the 144 kOhms did not jump into my face.
I can work around that problem. The deviation is linear, and I can calibrate that.
Johannes
Greetings Johannes
Message 13 of 13
(975 Views)