PARC 273 Galvanostat GPIB Commands
I need to read the measurement results from a PARC 273 Galvanostat/Potentiostat. Unfortunately, I do not have any documentation on the unit. Does anyone have a copy of the GPIB command set handbook which contains the protocol required to read the data from the unit? Can you please provide me with the commands and data format to transfer the result data to the PC?
Michael
Michael Munroe, CLD, CTD, MCP
Automate 1M+ VI Search, Sort and Edit operations with Property Inspector 5.1, now with a new Interactive Window Manager!
Now supports full project automation using one-click custom macros or CLI.
Message 1 of 17
(4,801 Views)
This is a way too late reply, but I have the same problem.
Please look at the following program as entered into the Headstart software. It causes the potentiostat to scan from -1 V to +1 V and then back to -1 V, collecting a total of 101 data points for each 100 milliseconds:
DCL
LP 100
SCV 2
MM 2
INITIAL 0 0
VERTEX 50 8000
VERTEX 100 0
BIAS -1000
INTRP 0
ASM
NC
CELL 1
TC; WCD; CELL 0
It was noted in the Headstart book that the last command in the sequence instructs the instrument to take the curve, wait until the curve is completed, and then turn the cell off.
Message 2 of 17
(4,461 Views)
Some more commands from the Headstart software:
  21
'CHRONOAMPEROMETRY EXPERIMENT
DCL        ' M273 DEFAULT PARAMETERS
MODE 2     ' POTENTIOSTAT
MR 2       ' 8000 COUNTS = 2 V ON MOD DAC
MM 2       ' ARBITR. WAVE. MODE
SCV 2      ' SOURCE CURVE #2
DCV 0      ' DESTINATION CURVE #0
PAM 0      ' NO AVERAGING
SIE 1      ' AMPEROMETRY  MEASURE CURRENT
INTRP 0    ' CLEAN STEPS
g 4
ASM        ' ASSEMBLE ARBITRARY WAVEFORM INTO SCV
NC         ' PRELOAD MOD DAC WITH FIRST POINT IN ARBITRARY WAVEFORM
P 5        ' PAUSE 5 SECS
CELL 1     ' TURN CELL ON
P 5        ' PAUSE 5 SECS
TC         ' TAKE CURVE
g 0
CELL 0     '  CELL OFF AFTER CURVE DONE
g 1
g 2
Message 3 of 17
(4,452 Views)
 32
AR 1             ' AUTORANGE CURRENT
AL -1            ' LOWEST CURRENT RANGE 100 mA
BIAS 0           ' POTENTIAL = 0V
MR 2             ' 8000 COUNTS = 2V ON MOD DAC
FP 0;LP 999      ' 1000 POINTS
TMB 1000         ' 1000 MICROSEC/POINT
S/P 1            ' ONE READING PER POINT
INITIAL 0 0      ' ZERO POTENTIAL FIRST
VERTEX 1 -2400   ' A STEP OF -0.6V
VERTEX 999 -2400 ' REMAIN AT -0.6V
ASM              ' ASSEMBLE ARBITR. WAVE. INTO SCV
PURGE 1;STIR 1   ' TURN STIRRER AND PURGE ON   1 = ON   0 = OFF
P 10             ' PAUSE 10 SECS
PURGE 0;STIR 0
P 1;DISP         ' DISPENSE A NEW DROP
P 5              ' PAUSE 5 SECS
P 5              ' PAUSE 5 SECS
USR1 TC;AL -7    ' PRECOMPILE THESE COMMANDS INTO USR1 FOR FASTEST RESPONSE
USR1             ' EXECUTE USR1
g 0
g 1
Message 4 of 17
(4,451 Views)
 19
'CHRONOPOTENTIOMETRY EXPERIMENT
MODE 1     ' GALVANOSTAT
MR 2       ' 8000 COUNTS = 2 X FULL SCALE ON MOD DAC
SIE 2      ' POTENTIOMETRY MEASURE POTENTIAL
g 5
ASM        '  ASSEMBLE ARBITRARY WAVEFORM INTO SOURCE CURVE
P 5        '  PAUSE 5 SECS AT CURRENT = 0
TC         '  TAKE CURVE
g 0
g 1
g 2
Message 5 of 17
(4,450 Views)
 31
I/E -4           ' 100 UA FS
BIAS 0           ' CURRENT = 0
TMB 200          ' 200 MICROSEC/POINT
INITIAL 0 0      ' ZERO CURRENT FIRST
VERTEX 1 -2000   ' A STEP OF 0.50 X FULL SCALE CURRENT RANGE
VERTEX 999 -2000 ' REMAIN AT 0.50 X FULL SCALE CURRENT RANGE
EGAIN 5          ' 2.000 V FULL SCALE POT'L MEASMENT
PURGE 0;STIR 0
P 5              ' PAUSE 5 SECS
TC               ' TAKE CURVE
g 0
g 1
Message 6 of 17
(4,449 Views)
 18
'CYCLIC VOLTAMMETRY EXPERIMENT FOR MODEL 253 VersaStat.
DCL         ' M253 DEFAULT PARAMETERS
SIE 1       ' COLLECT CURRENT
AR 0        ' AUTORANGE OFF
DCV 0       ' STORE DATA STARTING AT CURVE 0
MM 1        ' RAMP MODE
INTRP 1     ' TURN INTERPOLATION ON
MR 2        ' 8000 COUNTS = 2 V ON THE MOD DAC
g 3
TC          ' TAKE CURVE
g 0
CELL 0      ' TURN CELL OFF WHEN CURVE ACQUISITION IS COMPLETED
g 1
g 2
Message 7 of 17
(4,448 Views)
 19
DCL         ' PSTAT DEFAULT PARAMETERS
MM 1        ' RAMP MODE
g 3
P 5         ' PAUSE 5 SECS (NOT FOR VERSASTAT!)
TC          ' TAKE CURVE
g 0
g 1
g 2
Message 8 of 17
(4,447 Views)
 35
' CVCOM CYCLIC VOLTAMMETRY COMMAND FILE ADAPTED TO VersaStat(M253)
' THE M253 CAN NOT CONTROL A M303A DROPPING MERCURY ELECTRODE SO 303A
' RELATED COMMANDS HAVE BEEN REMOVED.
' CV EXPERIMENT 700 POINTS FROM -300 mV TO -1000 mV TO -300 mV  VS. Ag/AgCl
' NO. SWEEPS = 1 @ 100 mV/SEC
' CHARACTERS TO THE RIGHT OF THE APOSTROPHE ARE NOT TRANSMITTED TO THE
' POTENTIOSTAT.
I/E -3           ' 1 mA CURRENT RANGE
DCV 0            ' STORE DATA AT CURVE 0
SCV 2            ' SOURCE CURVE FOR ARBITRARY WAVEFORM
FP 0             ' FIRST DATA POINT 0
LP 700           ' LAST DATA POINT 700
TMB 500          ' TIMEBASE IN USEC
SWPS 1           ' SWEEPS = 1
S/P 41           ' SAMPLES PER POINT OF 41
BIAS -300        ' START EXP AT -300 mV
MM 2             ' ARBITRARY WAVEFORM MODE
INITIAL 0 0      ' FP 0  0 mV ADDED TO BIAS
VERTEX 350 -2800 ' POINT 350 -700 mV ADDED TO BIAS
VERTEX 700 0     ' POINT 700 0 mV ADDED TO BIAS
ASM              ' ASSEMBLE ARBITR. WAVE. INTO SCV 2
TC               ' TAKE CURVE
g 0
g 1
Message 9 of 17
(4,446 Views)
 38
' THIS COMMAND SET GIVES A REASONABLE CYCLIC VOLTAMMAGRAM FOR A 2.5mM
' CD(II) IN 0.1F KCL SOLUTION ON A M303A MEDIUM DROP
' POTENTIOSTAT.
P 5                 ' DISPLAY ABOVE MESSAGE FOR 5 SECS
IGAIN 10            ' INCREASE RESOLUTION BY 10
SWPS 1              ' SWEEPS = 1
INITIAL 0 0         ' FP = 0 &  0 mV ADDED TO BIAS
PURGE 1;P 5;PURGE 0 ' PURGE 5 SECS     NOT VALID FOR ORIGINAL VERSASTAT
DISP                ' MAKE A NEW DROP
P 5                 ' PAUSE 5 SECS
TC                  ' TAKE CURVE
g 0
g 1
Message 10 of 17
(4,445 Views)