Hi JamesC
LabVIEW 8.20 and the SVT 4.0, both are latest release.
Wav files that are from DAQmx 9233 are 50 kS/s, 24 bit mono 1sec. , 10 sec. and 30 sec. sample sizes have been tried.
Message Edited by Midlothian on 11-21-2006 08:48 AM
11-21-2006 10:07 AM
11-21-2006 10:35 AM
Thanks JamesC
That sounds promising. I attach a small 1sec sample from a calibrator set at 114db (ref 20mPa with 47.6 mV/Pa) an 1kHz.
Please remove or ignore the .jpg (it is added to permit forum upload).
11-22-2006 08:59 AM
JamesC,
Thanks for the pointer. I set the boolean for "ignore t0 continuity" to "true" leaving all the other checks and it now works correctly. The results are consistent with the calibrated signal; so all is well for the amplitude and frequency. I will leave this sub-vi in the edited state since it is unlikely to cause upset elsewhere. The ignore options could be "fed" from the higher level but this could change the behaviour of other toolkit VI's.
Cheers
11-22-2006 09:00 AM