Power supply controller Keithley 24xx via GPIB
Hello, I work first time with Keithley 24xx, I need to control power supply via labview. I ahve read a lot of exapmles according them i started to build my controller but there is some mistakes.
BR
Doni
(2,245 Views)
Hi Doniyor,
warm welcome to the LabVIEW forum!
Which "mistakes"? Any error messages?
It seems you are using some device driver VIs, but installed them in an unusual place.
It also seems you don't switch off the K24xx output after the loop as you wire a TRUE to "EnableOutput"…
(2,237 Views)
here mistakes:
Keithley 24XX.lvlib:Error Query.vi<ERR>
Instrument reports:
822,"Too small for sense range"
0,"No error"
<b>Complete call chain:</b>
Keithley 24XX.lvlib:Configure Output.vi
Voltage output.vi
(2,232 Views)
Hi Doni,
@Doniyor wrote:
Instrument reports: 822,"Too small for sense range"
This is an error occuring within your Keithley device. It is explained in the user manual for your device!
(As far as I remember Keithleys manuals even give suggestions to avoid those errors.)
From message I guess you need to select a different measurement range…
(2,228 Views)
Yesterday I modified and change a little bit but Keithley doesnt show datas which i supply
(2,165 Views)
Hi Doniyor,
you are using "German" regional settings (or a very similar regional setting) in your computer account, with a comma as decimal separator.
Those Keithley VIs are made for "English" regional settings, expecting a point as decimal separator.
The Keithley device follows the SCPI standard and also expects a point as decimal separator…
Do you see the problem?
Several options:
  • Use English numeric formats, with a point as decimal separator.
  • Set LabVIEW to NOT follow the OS settings, but stick to English numeric formats.
  • Edit the whole Keithley VI library to correctly format numeric values into strings as expected by the device.
(In my department I advise all colleagues to use English numeric formats to avoid any problems with hardware and other software. I only give support on software problems when they follow my suggestion…)
(2,145 Views)
There were changed smth. I supply 2V Keithley24XX shows everythings is right that what I supplied but after contact with element voltage change accidently.
I attached file. Please any advise
(2,101 Views)