Problem facing with Ni-DMM(PXI-4070)
Dear all,
      I faced a problem during the time of Acquisition for measuring the voltage signal (Range: 1mv to 15volts) through             NI-DMM (PXI-4070).That voltage was obtained form the UUT. The complete details of that problem are as             follows:
  1.) I measured the voltage signal Parallel with both NI-DMM Cards and Agilent DMM (through GPIB). While Acquiring the Data       sampling rate for the both NI and Agilent DMM'S are same (i.e., at 1 sample/sec).I acquired the data for 60 Sec for each       rotaion.For the Calculations Averaging the middle of the 60 sec data. The results coming from the UUT are not matching       with NI & Agilent .The obtained results from NI & Agilent DMM'S are shown in [DATA& Graph figure.]
  2.) I calculate the slope and intercept values of Average data and the rotations for both Ni & agilent DMM'S by using the        Linearity-Fit, Which was shown in the [DMM-DATA figure.].The slope values for both Ni & agilent DMM's are same but the        intercept values are different. Those calculations details are shown in[ DATA& Graph.]
  3.) Depending upon the intercept value and the voltage data coming from the DMM, the linearity curve will be               obtained. The linearity curves for the both NI & AGILENT DMM'S are shown in [Linearity Curves figure (3)],and linearity        values for different rotations are calculated by using the following formula:
                                    Linearity(L)= [(R*S+I)-D] / S ;
                Where R=Rotations ,S=slope,I= Intercept, D=DMM voltage value for that particular rotation.
                 The linearity values for different rotations are shown in [Linearity data figure].
   4.) While Acquiring the Voltage Data the following configuration is used
            DMM configuration
            Voltage range: 100 V
            Resolution   : 6 1/2
            Multi point configuration
            Sample count   : 0
            Multi point read
            Number to read : 2
            Maximum Time   : 1000 milli sec
(2,745 Views)
Looking at your data, I don't see anything unreasonable.
Between your Agilent and NI data, the worst difference in values between the two is 31.4 mV difference at the 30 rotation point.  Out of a 20 V range (+/- 10V), that is only 0.157% difference.
And for you intercept calculation, 8.09mV -6.95 mV = 1.14 mV which is only .0057% of the range.
Also, you say the DMM range is 100V.  If you calculate the errors based on that, then the error percentages are 1/5 the percentages stated above.
(2,718 Views)