Problem with NI 6221

Hi. I use one analog output channel and 5 input channels. They all are scanned. When I work only with one output and input chanels (4 channels are not connected), I can see on them the same signal as on working channel.
I read the following issues:
http://digital.ni.com/public.nsf/websearch/B9BCDFD960C06B9186256A37007490CD?OpenDocument
http://digital.ni.com/public.nsf/websearch/4E9B8B3F8FD3A31B862562B5004EA1D9?OpenDocument

I want to understand, what are the solutions of my problem?

What is the source impedance - выходное сопротивление источника?

Thank You.
0 Kudos
Message 1 of 9
(4,026 Views)
I forgot. I use RSE (not differential inputs). How are they connected in NI 6221?
May be I have to connect the second input from two to ground?
0 Kudos
Message 2 of 9
(4,025 Views)
This is normal behaviour. The 6221 has a multiplexer witch switches the input channels of the AD converter.
A residual voltage will be present at a floating input.
Ground non-connected inputs and you will be OK

Ton
Free Code Capture Tool! Version 2.1.3 with comments, web-upload, back-save and snippets!
Nederlandse LabVIEW user groep www.lvug.nl
My LabVIEW Ideas

LabVIEW, programming like it should be!
0 Kudos
Message 3 of 9
(4,022 Views)
You mean, that I have to connect with ground other 11 input channels. (5 channels are connected allways and I work with them all but sometimes I work only with one or two channels)
0 Kudos
Message 4 of 9
(4,018 Views)
Floating inputs are bad inputs!
But I usually only ground the first floating input, that will do most of the time.

Ton
Free Code Capture Tool! Version 2.1.3 with comments, web-upload, back-save and snippets!
Nederlandse LabVIEW user groep www.lvug.nl
My LabVIEW Ideas

LabVIEW, programming like it should be!
0 Kudos
Message 5 of 9
(4,016 Views)
The decision is not to use floating channels?

floating channel - sometimes connected channel, sometimes disconnected?


0 Kudos
Message 6 of 9
(4,006 Views)
A floating channel is a channel that is not connected, just ground it when you pull of the signal.

Ton
Free Code Capture Tool! Version 2.1.3 with comments, web-upload, back-save and snippets!
Nederlandse LabVIEW user groep www.lvug.nl
My LabVIEW Ideas

LabVIEW, programming like it should be!
0 Kudos
Message 7 of 9
(4,002 Views)
Are these floating channels set in the scanning list?
0 Kudos
Message 8 of 9
(3,979 Views)
No, I mean phisical, if you remove the wire from your connection board, you have a floating input, connect this input to the ground via a terminator or something.
That is enough.

Ton
Free Code Capture Tool! Version 2.1.3 with comments, web-upload, back-save and snippets!
Nederlandse LabVIEW user groep www.lvug.nl
My LabVIEW Ideas

LabVIEW, programming like it should be!
0 Kudos
Message 9 of 9