Problem with SCPI commands for Fluke 8846A
Hey guys,
I am trying to access an Fluke 8846A (https://download.flukecal.com/pub/literature/8845A___pmeng0300.pdf) via GPIB/VISA. However, many SCPI commands are not recognized, and I can not figure out why. For example, "*IDN?" will give me the response as described in the manual, "*RST" also works fine. When I try other commands, like "MEAS:volt:ac? 10", which is listed as an example in the manual, the device is showing errors -102 (Syntax Error) and -420 (unterminated command).
Do you have an idea on how I can solve this problem?
Kind regards
(3,192 Views)
@verfelixt wrote:
Hey guys,
Kind regards
Well yes.
To set a 10A range for AC Voltage Measurement use
"MEAS:volt:ac 10" no question mark.
To Query the configured AC Voltage Measurement range use
"MEAS:volt:ac?" The device will respond with 10 or whatever range is actually set.
You have simply misunderstood what you read in the manual.   THANK YOU for actually reading it 😊
(3,186 Views)
That does not work, too. As described, "*IDN?" or "*RST" are working, commands like "CONF?", "MEAS:volt:ac?"/"MEAS:volt:dc?" will not work and cause the described errors on the device. Same goes for other commands that do not prompt a response from the device like "MEAS:volt:ac 10" (no question mark 😉 )
(3,178 Views)
Accepted by topic author verfelixt
The prototype is
MEAS:VOLT:AC <range>, <resolution>
Range can be "min", "max" or some value less than the devices maximum.
To set just the range use
MEAS:VOLT:AC:RANG <VAL>
NOW, that's assuming the language is set to l1
In l2 you would send
VAC
RANGE <ENUM from table 23a or 23b>
In l3 you would need a urinalysis screening prior to reporting to work. But get and out functions were very time saving back when communicating with smoke signal speeds.
(3,147 Views)
Thank you very much, the wrong language was the problem, it was set to L2, not L1. I should have read the manual more properly 🙂
(3,117 Views)
No, thank you for linking it!  I've used similar devices in the past and that darn manual was probably one of the least clear I have read.  No shame for a novice to come away with a headache after reading that!  Glad I could help.
(3,107 Views)
You did your homework by reading the manual and tried your best to implement it.  No shame in that!!!  🙂
Bill
CLD
(Mid-Level minion.)
(3,098 Views)