no bother, I like the feeling that my time developing pH electrodes wasn't totally wasted.....
Yes, Sensorex. I remember testing them in the lab a few years ago. They work on an ISFET basis. Nice idea in itself, not requiring a glass membrane, but their reference electrode junctions are really really bad. They have solidified gel electrolytes which cet diluted rather quickly, and ceramic pin diaphragms which get blocked really quickly. All in all, the reference electrode simply isn't very good.
If I may do some advertising for an electrode I helped develop, look
here. I would recommend the pH+. It also has a gellified reference electrolyte, but it has plenty of KCl crystallized within to help replenish anything which would otherwise dilute the electrolyte. In addition, it has a ground sleeve ring diaphragm which is simply miles ahead of ceramic pin diaphragms in any kind of bio application. It won't clog (well sometimes it will, but not until LONG afgter ceramic pin ones already have).
Errors you need to look out for are
1) Alkali error : non-linearity of the pH glass above pH 12. If autoclaving the electrode is neccessary, this can become a problem at pH 10 eventually as the glass deteriorates.
2) Diffusion potential : The liquid junction of the reference electrode should theoretically have zero potential drop. This is, of course, never actually the case but gellified reference systems (especially ceramic pin ones) can have junction potentials of up to 0.5 pH units! Ring type liquid junctions are much superior in this regard (0.2 pH max IIRC).
3) Stirring error : Often confused with the Diffusion potential, the difference in measured value when stirring or not stirring. In case of doubt, the value WITHOUT stirring is correct. Ring diaphragms are here again far superior. Ceramic - 0.2pH, Ring - 0.05 pH.
4) Increasing membrane impedance : If autoclaving glass pH electrodes, the impedance of the glass gradually rises. This can eventually become a problem above 1-2 Gigaohm membrane resistance. ISFETS don't have this problem, but they have lots of others (Long-time drift being one big one).
I think you'll find that the robustness difference between your linked Sensorex electrode and a proper Bio pH-electrode (as the pH+ linked above) is surprisingly small. Sensorex argues that HF has no affect on their electrodes. We tried this out once and it died after the same exposure to HF as the pH+ electrode. The fact that the body is made of glass-reinforced plastic doesn't help.
All in all, my experience has been that ISFETs are a nice idea, but the people developing them are electronics people who simply have no clue about the physics of pH measurement. The electrodes are designed accordingly. I think you'll be better off with a glass electrode, assuming you can sort out the impedance problem...... That is the only real advantage of an ISFET - low impedance - in the range of Kiloohms.