Rise rate calculation
I’m trying to calculate rise rate for a pressure measurement. It is defined as in the picture (riseratecalc.png). Up to t0 there is no applied pressure. After t0 there is an applied voltage and it increases. t1 is usually defined as t0 plus 1 seconds. Thus if I find t0 , I can find t1 and P0 and P1 (corresponding pressure values respectively). But there is a difficulty in finding t0 instant. When there is no pressure in the line, there is a very small, varying voltage (+3,+5,-3,-5 mV). (With my pressure calculation code I will calculate this varying voltage emprically then use this as threshold voltage. ) In order to accomplish this, I wrote a small code (riserate.vi) I created a voltage array with 10 elements and compared its elements with a threshold voltage. If it is equal or greater than this threshold voltage I break the while loop, grap this index (it shows t0 instant) and use this index number to find P0, then add a sampling ratio * t seconds to find t1 and use this index to find P1. Then I substract P0 from P1 and calculate rise rate. My problem is how to integrate this to my pressure reading code (press1.vi). I measure pressure from two channel, but I will use rise rate calculation only in outlet channel. I read the voltage values after mean.vi and pressure values after making scaling. It is probably more clear what I want to do in press2.vi. There are also some more issues. I run press.vi , it is working fine , measuring the pressure correctly. But when I run it highlight execution mode, it gives -200279 error. I read some messages related to this error, tried to solve playing with sampling rate and buffer size but couldn’t manage. I also didn’t understand why I get this error only in the normal mode. It is very annoying now to deal with this error while I try to analyze and fix the other problem. Finally, I put a over limit alarm in press1.vi, I took it from NI examples (Temperature System Demo) . But in that example it was hidden when there is no over limit. It only comes to display during over temp. But in my code it is visible all the time. I want to hide it somehow.
Meanwhile, I want to say that, there are very helpful people in this forum. I didn’t meet any replies in the other forums as quick and reliable as here. Thanks for your concern.
Ahmet
(4,240 Views)
Hi Ahmet,
It can cause problems when you run highlight execution when you are doing continuous data acquisition. You have it set up to acquire at a specific rate, and even though you are doing highlight execution, you are still going to acquire at that rate. The problem is the rest of the program is occuring much slower, so you are not pulling samples out of the buffer fast enough and you get a buffer overrun error.
As for the over limit, I looked at that example and what they did there was set that boolean to be transparent when it was off.  You can set that by right clicking on it and going to properties, appearance tab. Click in the color box for off and then in the upper right corner you will see a box with a T in it. That stands for transparent. I hope this helps!
Nick R
NI
(4,181 Views)
Hi Nick R
Thanks for your reply. But let me change my explanation about error -200279 in this way. Even though I use Cont Acq&Graph Voltage-Int Clk.vi example I get this error during highlight execution mode. Therefore I think the cause of this error can be different.  My system is capable enough, I think (Intel Core Duo 2 GHz, 1 GB RAM) . I have examined all the messages related to this code, but I didn't see anyone complaining about getting the error only in highlight execution mode.
The other problem was solved when I set that boolean to be transparent when it was off.
Ahmet
(4,150 Views)
Ahmet,
Just to confirm that I am correct, I ran that example in highlight execution and got the same error you are getting. It really has nothing to do with the specs of your computer, it has to do with the buffer size of your DAQ card. For example, I have a PCI 6251 card in my computer, and its analog input buffer is 4095 samples. That is the maximum number of samples that will fit in the buffer. So if you are running that example at the default rate (10000 Hz), that buffer will be filled in under .5 seconds if you are not immediately taking samples out of the buffer.
In this example, as soon as the DAQmx Start executes, the card starts acquiring samples and places them in the buffer. The samples wait in the buffer until DAQmx Read executes. This is not a problem when running in normal mode because the read happens immediately after the start, so the buffer does not get full. When you run in highlight execution, there is a long delay between the start and the read, significantly more than .5 seconds, so you get this error indicating that there has been a buffer overflow. When the buffer gets full the oldest values in the buffer are overwritten by the new values, as indicated in the error message you are getting. Let me know if you need any further clarification.
Nick R
NI
(4,131 Views)
Hi Nick,
I got what you mean, but I still need some clarification 🙂 Actually I use USB 6009 as hardware, but I use PCI 6221 virtually to simulate my programme. I've checked PCI 6221 specifications and found that its input FIFO size is also 4095 samples. My real card (USB 6009) has 512 bytes AI FIFO (which means 4095 samples). But what is the meaning of max. sampling rate in the specifications ( which is 48 kS/s for USB 6009, 250 kS/s for NI 6221) . In DAQmx Timing.vi we determine the sampling rate and samples per channel (buffer size for continuous acquisition). Thus I think 48 kS/s (for USB 6009) and 4095 samples are max. vaules for sampling rate and buffer size respectively. Am I right? Your comment also shows that I can not sometimes use highlight execution mode for code analyse effectively.
And I have another question, maybe I have to send it seperately with another message subject. I want to read some voltage values from a txt file (test.txt) and insert these values to an 1D array. I want the first value to be first element of the array, second value to be second element and so on.
Ahmet
(4,115 Views)
Hi Ahmet,
The buffer size and the samples per channel are two seperate things. Also there are two different buffers at play here. There is the buffer on your daq card, 4095 samples in this case, and the buffer DAQmx allocates on your computer. Information is passed from the card's buffer to the computer's buffer seamlessly, you don't have to worry about that. Samples per channel is the number of samples DAQmx pulls from the buffer for each channel each time the DAQmx Read is called.
So the situation is that DAQmx allocates this buffer on your computer, whatever that may be, it is a finite size. So if you start your task and then have a long delay before you start reading values out of the buffer, then your buffer will get full and overflow and you get that error.
To specifically answer your questions, yes the 48 kS/s and the 4095 samples are the max values, but that buffer is the buffer on the card. DAQmx automatically allocates the buffer on your computer based on your rate and samples per channel. Also, you are correct in saying highlight execution is not as effective to analyzing problems when you are acquiring data. For data acquisition you have a defined rate that you want to acquire that data, but when you use highlight execution the whole program gets slowed down, so there will be problems.
For your last question, if you have the voltage values in a tab delimited text file then you can just use the Read from Spreadsheet File.vi to read the values directly into an array.
Nick R
NI
(4,104 Views)