Sampling Rate Question

Hello,

 

I have been using an NI cDAQ-9188 with 2 NI 9324 modules to collect data. Using LabVIEW 2012 SignalExpress under DAQmx AcquireI I set the sampling rate to 250Hz and the samples to read to be 250.
When I output the logs to an ASCII file it says the dt=0.000605. With 250Hz I would be expecting 0.004. Any idea what the cause might be?

0 Kudos
Message 1 of 4
(2,958 Views)

Hi Jordan, I'm not familiar with SignalExpress, could you post some screenshots of what you are doing?

0 Kudos
Message 2 of 4
(2,934 Views)

Hi Jordan,

 

when in trouble you should read the manuals!

 

The samplerate given on page 6 fits to your 0.6ms dt value…

Best regards,
GerdW


using LV2016/2019/2021 on Win10/11+cRIO, TestStand2016/2019
0 Kudos
Message 3 of 4
(2,926 Views)

As GerdW was getting at, the slowest rate that the 9324 can sample at is 1.65kS/s/CH.