04-18-2017 10:55 AM
Hello,
I have been using an NI cDAQ-9188 with 2 NI 9324 modules to collect data. Using LabVIEW 2012 SignalExpress under DAQmx AcquireI I set the sampling rate to 250Hz and the samples to read to be 250.
When I output the logs to an ASCII file it says the dt=0.000605. With 250Hz I would be expecting 0.004. Any idea what the cause might be?
04-18-2017 11:29 AM
Hi Jordan, I'm not familiar with SignalExpress, could you post some screenshots of what you are doing?
04-18-2017
11:46 AM
- last edited on
09-26-2025
12:57 PM
by
Content Cleaner
Hi Jordan,
when in trouble you should read the manuals!
The samplerate given on page 6 fits to your 0.6ms dt value…
04-18-2017 12:00 PM
As GerdW was getting at, the slowest rate that the 9324 can sample at is 1.65kS/s/CH.