Scaled physical units for a loadcell
Hi!
I have a load cell that i want to measure the weight of an object in the unit gram (g). I have a Ni 9237 connected with my load cell. I'm aware that the NI 9237 displays the signal in mV/V. I want to scale this unit to gram. In my labview code I use the DAQ assistans and i know that I have to set my specific values from my load cell certificate in here. But i don't really see where all of my values from my load cell should be inserted. I will attach a picture of my certificate if that will help you understand it better.
I have tried a custom scaling with scaling parameters. The values for slope and intercept I had to calculate by measureing wieghts of 10g 20g 30g and so on , to see what values of mV/V they generated. Then i made a regression and took that slope and intercept, but that didn't work. I gett errors like, 200077" Requested value is not a supported value for this property. The property value may be invalid because it conflicts with another property", and it says that I need to select Newtons instead of mV/V.
I have searched on the forum, but I haven't found a direct guide to this. I know there is but seems that I'm not very good at searching info.
To be noted I have the student edition of Labview 2013
Best Regards Maurlind
Message 1 of 9
(5,535 Views)
Hi maurlind,
that certificate doesn't provide the details of relation between weight and output. All you see is dependence of output signal vs. supply voltage (as a typical value of full/wheatstone resistor bridges).
For your error message: we cannot help as long as you don't attach your VI. (Saved for LV2011 would be fine for me.)
Which polynomial regression values did you use?
Ah, I see you're using that resistor bridge module. What excitation voltage do you use?
What's wrong with Newton outputs? You are measuring a force (aka load)!?
Message 2 of 9
(5,532 Views)
Hi!
I use 10V as excitation voltage. There is nothing wrong with newtons as outputs, I can always convert it later in my code. I tried to change to it but it didn't' work.
I used a linear regression withe the formula y=0,0362x-0,0187 where x is in gram and y in mV/V. the values 0,0362 and -0,0187 where inputted in the custom scaling parameters. Se the attached print screens. For max I inputted 3000 which is the maximum load for my load cell and min -3000, in grams.
My error comes from the DAQ assistans when I'm done with my settings and want to exit. It has nothing to do with my code.
Message 3 of 9
(5,520 Views)
which output values do you get without applying a custom scale?
Message 4 of 9
(5,512 Views)
Hi!
With output values do you mean the values from the signal?. It is calibrated so it should give out zero, or around it offcourse.
With out applying a custom scale I use the default setting except for the vex value.
Best regards Maurlind
Message 5 of 9
(5,501 Views)
Hi maurlind,
Did you read the comments on V_excitation in the manual of your NI9237?
Yes: which signal value do you get at 3kg, at 2kg, at 1kg, at zero load?
Message 6 of 9
(5,494 Views)
Hi!
I have read about the NI 9237 from this link http://www.ni.com/pdf/manuals/374186e.pdf
What I understand is that the NI 9237 uses an internal voltage , 2.5, 3.3, 5 and 10V since my load cell need 5-12V according to my certificate I can use 10V from the NI 9237 and use then use the Vex value as 10V, or I'm I wrong? I don't know what exact voltage I must use or how I can adjust it since I haven't found any info of it. I don't have any physical manuals for my instruments at least I'm aware of, maybe in a box somewhere in my Lab.
Here are the output values that I have noted:
g mV/V
10 0,35
20 0,7
30 1,05
40 1,425
50 1,8
60 2,16
70 2,52
80 2,875
90 3,22
100 3,6
1000 35,28
2000 70,53
3000 105,78
Message 7 of 9
(5,483 Views)
Accepted by topic author maurlind
Hi maurlind,
What I understand is that the NI 9237 uses an internal voltage…
That manual also mentions some power limits. Did you read the manual?
Here are the output values that I have noted…
The output signal of a bridge is of unit "mV" - not "mV/V"!
You get about 35µV/g (with current excitation voltage). All you need to do is to multiply your signal with this number (signal * 1g/35µV) to get your load. Either create a DAQmxScale to do this or use simple math on your signal…
Message 8 of 9
(5,479 Views)
Hi !
It worked with multiplying with that factor.
I also read that the power limit is at 150 mW. According to the pdf file i can use a full bridge of 350 ohm with 3.3V. Then I should set the Vex value to 3.3V ? Don't i have to set my NI 9237 to the same voltage as well, if so how do you do that, didn't read anything about that in the manual.
Best Regards Marulind
Message 9 of 9
(5,475 Views)