Serial read/ write primitive slow in 2016 compared to 2012
Thanks for the response. I failed tro mention the baud rate of 460800. The 20 ms is NOT between command / response (which is completed in less than 4 ms) but between consecutive write/read sequences. This seems to indicate that the HW is able to handle the rate, unless the 20ms is needed for the HW to re-configure from read to write mode. I'll try a test case in full duplex mode where the TX is wrapped back to the RX lines to observe if the delay is maintained.
Message 11 of 13
(1,066 Views)
If you're using a USB-to-serial adapter, the delay may be introduced by that. FTDI chips, for example, are by default configured to wait for a certain amount of time after receiving any bytes before transferring them to the PC. It looks like the default setting is 16 ms. Here are a few links:
http://www.ftdichip.com/Support/Knowledgebase/index.html?an232b_04adjlatency.htm
http://www.ftdichip.com/Documents/AppNotes/AN232B-04_DataLatencyFlow.pdf
Message 12 of 13
(1,059 Views)
That solved it! The "Latency Timer" in the port properties (Advanced) setting was set to 16 ms. Changed it to 1 ms and things are screaming now. Thanks for sharing your experience, it would have taken me a while to get to there. Case closed and hopefully this will help someone else in the fuiture as well.
Message 13 of 13
(1,056 Views)