Setting IP address on PXI-8196 Embedded Controller
I have a PXI-8196 Embedded Controller that has been working, but recently failed to boot properly. I formatted the controller harddrive, and followed the directions given in
https://knowledge.ni.com/KnowledgeArticleDetails?id=kA00Z0000019LSXSA2&l=en-US
The IP address shown in NI-MAX and on the controller video output is 0.0.0.0. I have tried everything I can think of to change this address, but have not been successful. Does anyone have any suggestions beyond what is given in the NI link?
The controller didn't keep the IP address even though it was configured to keep the address settings when it was formatted. I've tried the physical DIP switches, tried a direct ethernet connection, and tried the USB boot drive method. Host machine firewall is off. Tried various host machine ethernet adapter settings. Norhing seems to work. When using NI-MAX to change the address, no errors are reported, and it reboots the controller, staying with 0.0.0.0.
(4,583 Views)
Drat!  I know I've done this, but I haven't worked with a PXI system for more than a decade, so my memory is fuzzy.  If you set it up with an IP of 0,0,0,0 (which, if memory serves, can be done by using the four little switches located inside somewhere, see the helpful figure in the Manual -- you might also be able to set it from the BIOS), will MAX allow you (Networking tab?) to set the IP?  I know I wanted to use a 10.x.x.x IP and have a second controller in my PC that was directly wired to the PXI using the same IP range, which worked very nicely ...
Bob Schor
(4,540 Views)
Hi Bob, in MAX I can see the controller listed with IP address 0.0.0.0, and when I plug a display into the controller itself, that also shows 0.0.0.0. When I use MAX to change the IP address and then press "Save" at the top, it re-boots the controller and still shows 0.0.0.0. I remember fighting with this a few years ago and had found a way to set the IP address, but don't remember the details. The BIOS on the controller only allows clearing the IP address to 0.0.0.0, and doesn't have the capability to specify a particular address. MAX does have the ability to create a USB boot drive; I've tried that without success.
(4,525 Views)
Is the BIOS Date/Time incorrect after a power down? Perhaps if the backup battery is dead it could be affecting the IP address setting?
Just a WAG, I have zero experience with this hardware.
-AK2DM
(4,517 Views)
Good point about the battery; I'll check it out.
I did find another link that may be relevant:
Using a Utility USB Drive to Format my Real-Time Phar Lap ETS PXI Controller - NI
so I will try that. In particular, I will try the Windows/Other OS setting. It may be possible to set the IP address under Advanced Ethernet Options.
(4,512 Views)
Accepted by topic author Chris_12345
I got it working! Here are the steps that I took:
Controller has direct ethernet to host machine
Controller has new battery (type 2032)
Insert a small (256 MB) USB drive into the host machine
NI-MAX > Tools > Create Desktop PC Utility USB Drive
Insert USB drive into PXI-8196 controller
Boot controller while pressing Delete key to enter BIOS
BIOS > LabVIEW RT > Boot Configuration > Windows/Other OS
BIOS > LabVIEW RT > Reset IP Address: YES
BIOS > LabVIEW RT > Disable Startup VI: YES
BIOS > Boot > Boot priority order: USB first
BIOS > Exit › Exit Saving Changes
In boot utility, select 3. Reset network configuration
In boot utility, select 2. Boot into safe mode
Wait for display at top to show new IP address: 169.254.168.214
Host machine: Control Panel \Network and Internet Network Connections ›
Ethernet › Properties › Internet Protocol Version 4 > Properties
Set IP address 169.254.168.100, subnet mask 255.255.255.0
Close and re-open NI-MAX (important)
NI-MAX > Remote Systems › "Controller" › Tab: Network Settings
Configure IPv4 Address: Static, 169.254.168.214, 255.255.255.0
NI-MAX > Remote Systems > "Controller" > Software
Select items to install software
(4,446 Views)
THANK YOU for not only solving your own problem, but so carefully documenting the steps you took "for the next person with the same issue".  Your choice of Message Topic is also excellent!  I gave you a Kudo!
Bob Schor
(4,423 Views)