Hi all,
I'm using NIMAX with the following simulated hardware:
I'm trying to run a very very basic code that still captures the problem (attached), but i get the following error when I press the "stop" button on the front panel.
This doesn't seem to make any sense, as each 9205 module should be able to handle 16 differential measurements for a total of 32 differential measurement. I even tried switching the simulated module positions around, wondering if it was a limitation with the cDaq9174. Nope, still present. Also, it seems that only the "voltage2" is afffected. It does not throw an error if I set the ai range from 0 to 7. It starts happening when i set the ai range from 0 to 8.
I do not yet have the hardware on hand to test this out without simulated hardware, but I would very much like to get to the bottom of this.
Thanks in advance!
05-28-2013 05:51 PM
Update: Still no go, but it seems to resolve itself when I set it as "RSE" or "NSRE." Unfortunately, I need differential inputs, but it's interesting that it resolves itself when I change that setting. Does this spark any ideas in anyone?
Any/all help would be greatly appreciated!
05-28-2013 08:14 PM
ARGH Bloody hell. It took me a while, but I found the answer.... So stupid!
I'm writing down what happened just in case anyone else was as stupid as I was. Unlikely, but who knows?
The Differential channels are not paired continuously and cannot be called as such. In other words, you cannot simply call ai0:15. It is actually split into ai0:7, and ai16:23.